sci_oxy3835_wphase_oxygen
is a direct output from the optode. Since no information is fed from the glider into the optode, this is a non-compensated value.
There is no difference to sci_oxy3835_oxygen apart from a different setting internal to the optode (wphase spits out more information).

When trying to reproduce the calculation of the optode (convert phase(s)&temp to oxygen) you however need some internal settings of the optode, which the glider and the glider's data files are unaware of !!! That is e.g. the internal salinity setting of the optode. I do not recall what the recommendation was. Either 0 or 35 PSU. Try using the either of these values. Or maybe somebody tried to be very s…