For each temperature run should we leave all materials, reagents and sensors in the lab/workshop/room overnight (or as recommended for 8 hours). When time is limited, is it ok to begin calibration when the sensor equilibrates (if this occurs in a time period shorter than 8 hours)? |
Beta Was this translation helpful? Give feedback.
Replies: 2 comments
-
There are two things here. The other part, and perhaps more important, is that the foils need to be fully hydrated, which does appear to take some time, at least 1 hour if it was dry to begin with. It is probably fine to do the calibration after one hour if the sensor was kept wet, but I have not actually tested it and generally I put the sensors in a bucket overnight prior to calibrating. |
Beta Was this translation helpful? Give feedback.
-
Agree with what Tom is writing. Hope the pre deployment calibration went well. |
Beta Was this translation helpful? Give feedback.
There are two things here.
The general idea is that if your instrument and gear is warming/cooling rapidly it's difficult to get a good calibration. I've found though that unless you're in a temperature controlled room (not a bad idea) you're going to get temperature fluctuations during the day, I would say that if the sensor (both oxygen and temperature) is stable over 10 minutes it would be fine.